Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer E LOW Water Message |
Problem Description:
Tuttnauer E LOW Water Message |
Reasons |
1. Tuttnauer E LOW Water Message This message will be displayed if during a normal heat up stage the system determines that there is insufficient water in the chamber to complete the cycle. This determination is made by the combined Input of two (2) sensors: The water electrode and the safety thermostat Also if a power failure occurs during the heat or sterilization stage after the pwoer returns the system will check the water electrode to see if there is sufficient water in the chamber in order to resume the cycle. If not the cycle will be aborted, the message LOW Water will be displayed, and the cycle fail indicator will light. Possible cause: 1) Insufficient water in the chamber at beginning 2) Leakage somewhere 3) Power off and back |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |