Tuttnauer E/EA/EK/EKA LOW PRES
Problem Description:
Tuttnauer E/EA/EK/EKA LOW PRES
Reasons
1. Tuttnauer E/EA/EK/EKA LOW PRES

This message is displayed, the Cycle Fail indicator lights, and the cycle is aborted if the pressure drops 4 PSI (0.27Bar) below the required sterilizatin pressure

Possible causes:

1) SSR
2) Insufficient water inside chamber
3) Heating Elements can not on again (SSR, or Control circuit)
4) Bad Heating elements (may be one or 2 damaged)
5) Safety thermostat for T93N6 and T93N5
6) bad pressure sensor

The technical information is just to help sterilizer users as reference to do their trouble shooting.