Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer E/EA/EK/EKA HIGH TEMP |
Problem Description:
Tuttnauer E/EA/EK/EKA HIGH TEMP |
Reasons |
1. Tuttnauer E/EA/EK/EKA HIGH TEMP This message is displayed, the Cycle Fail indicator lights and the cycle is aborted if the temperature rises 9F (5C) above the required sterilization temperature during the Sterilization phase of the cycle This message will also be displayed if the Temperature Sensor is damaged. In this case the message will appear just before the heat phase starts Possible causes: 1) The heating elements are remaining on instead of on/off, check the SSR 2) The msssage can also indicate a bad temperature sensor: the massage will display any time during the Heat Up phase |
|
|||||