Sterilizer Manufacturer: Midmark |
Model of Sterilizer: M11 |
Problem Title:
ritter temp adjustment |
Problem Description:
ritter speedclave wont go past 260 degrees how do i adjust it to 270? |
Reasons |
1. Temp/Pressure gage may not be reading correctly. Bellows may not be operating correctly. Relay may need to be calibrated. |
Solutions |
1. If bellows and temp/pressure gage are good, then adjust the relay by loosening the setscrew in the mount that holds the relay to the side of the chamber, turn screw next to setscrew clockwise to increase temp and tighten setscrew so temp rise stops at 270 or before safety starts to release. |
|
|||||