Reasons |
1. SCICAN Statim Error Code #1 The message means: the cassette temperature failed to reach 95C in three minutes possible reasons: 1) Boiler does not heat up. No power to boilder. Service required 2) Check the thermal fuse. Using ohmmeter to test J1-3 & lower terminal on boiler. Ohm should be less than 1 ohm. If not a service required 3) Check teh resistance reading between J1-3 and J1-4 should be 11 ohms approx. If not technical service required 4) Check the voltage. If not, shop repair needed. 5) Check voltage at boiler terminals. |
|
|||||