Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: Statim 2000 |
Problem Title:
SCICAN Statim error code 10 |
Problem Description:
SCICAN Statim error code 10 |
Reasons |
1. SCICAN Statim error code: 10 The cassette temperature has fail to drop to 115C during the upwrapped or wrapped cycle or the temperature has failed to drop to 110C during the rubber and plastic cycle in the purge conditioning stage 1) Check for kinked or pinched exhaust tubing 2) Check for a clogged venturi in the left rear of the cassette tray. Clean as needed 3) Solenoid valve is failing to open. Disassemble and check it. 4) Check for power to solenoid |
|
|||||