Problem Title:
SCICAN Statim 2000 error code 11
Problem Description:
SCICAN Statim 2000 error code 11
Reasons
1. SCICAN Statim error code: 11

The cassette temperature has failed to drop to 102C within 60 seconds of the end of a cycle during venting

1) Check for kicked or pinched exhausting tubing
2) Check for a clogged venturi in the left rear of the cassette tray
3) Solenoid valve is failing to open, disassemble and check the solenoid
4) Check for the power to the solenoid

The technical information is just to help sterilizer users as reference to do their trouble shooting.