SCICAN Statim error code: 15
Problem Description:
SCICAN Statim error code: 15
Reasons
1. SCICAN Statim error code: 15

The cassette temperature is 3C or more above a set point during the sterilization phase of the cycle

1) Check for kinked or pinched exhaust tubing
2) Check for a clogged venturi in the left rear of the cassette tray
3) Solenoid valve is failing to open, disassemble and check it.

The technical information is just to help sterilizer users as reference to do their trouble shooting.