Sterilizer Manufacturer: Market Forge |
Model of Sterilizer: Market Forge STM-E |
Problem Title:
How to contain steam in the Market Forge? |
Problem Description:
Is a way to capture the steam and to send it to the outside of the room? |
Solutions |
1. Here are the two options available. OUTSIDE VENTING Connect 1/2" (13mm) nominal tubing exhaust to outside vent connection located on the top of the control housing. IMPORTANT: Exhaust line must be vented to the outside to eliminate the exhausted steam and the accompanying noise from entering the room. Use 1/2" (13mm) copper tubing or suitable alternate. Length of the line should not exceed 15 feet (4.5 meters) and should have a minimum of bends. The line should slope downward after leaving the sterilizer in order to insure condensate drainage. WATER-COOLED EXHAUST CONDENSER If outside venting is not possible, an optional water-cooled condenser is available for connection to an open drain. If required order part number 95-0436 kit. |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |