SCICAN Statim error code: 25
Problem Description:
SCICAN Statim error code: 25
Reasons
1. SCICAN Statim error code: 25

The software has failed to detect a need to pump water in 90 seconds.

1) Boilder does not heat up
2) Check thermal fuse
3) Check resistance reading between J1-3 & J1-4. Should be 11 ohms approximately
4) Check the line voltage at J1-3 and J1-4 during warming up
5) Check for line voltage at boiler terminals

The technical information is just to help sterilizer users as reference to do their trouble shooting.