Sterilizer Manufacturer: Prestige |
Model of Sterilizer: |
Problem Title:
Prestige 2100 Vacuum Error Code E02 |
Problem Description:
Prestige 2100 Vacuum Error Code E02 |
Reasons |
1. Prestige 2100 Vacuum Error Code E02 The autoclave is fitted with an electronic comparator that checks the following condition during the sterilizing cycle t E02 = sterilizing hold period fault d E02 = PT100 boiler and chamber probles do not agree b E02 = Temperatures outside of control band Autoclave has failed to sterilize. 1) Check that the value used in calibration modes CAL01 and CAL02 are as per specification. 2) Check chamber probe is correctly inserted 3) Check that both PT100 probes have a similar resistance 4) Check all pipe work for leaks 5) Power control module fault |
|
|||||