Sterilizer Manufacturer: Pelton & Crane |
Model of Sterilizer: |
Problem Title:
Delta 8/10 Water flows into bottom of Sterilizer when unit is not in FILL cycle |
Problem Description:
Delta 8/10 Water flows into bottom of Sterilizer when unit is not in FILL cycle |
Reasons |
1. DELTA 8 10 Water flows into bottom of Sterilizer when unit is not in FILL cycle Reasons: 1) Leaking fill solenoid valve 2) Hole in condensing tube 3) It is normal, each time turn on the power, there is a little water coming in chamber Solutions: 1) Clean or reaplce fill solenoid valve 2) Replace the condensing tube |
|
|||||