Sterilizer Manufacturer: Pelton & Crane |
Model of Sterilizer: |
Problem Title:
DELTA 8 10 Error P2: Sterilizer Pressure reaches 241 Kpa before 135C temperature |
Problem Description:
DELTA 8 10 Error P2: Sterilizer Pressure reaches 241 Kpa before 135C temperature |
Reasons |
1. DELTA 8 10 Error P2: Sterilizer Pressure reaches 241 Kpa before 135C temperature First, check the sterilizer can return the air/steam back from chamber to reservior or not between 90C to 110C. It is very important to make sure all air will be purged out of the chamber. Only pure steam, the temperature can match pressure. If not return, check the bellow solenoid or the Power PCB board |
2. If the air purging works well between 90C - 110C, then check the RTC (steam) sensor to make sure it is properly conntected to the Power PCB board