Sterilizer Manufacturer: Pelton & Crane |
Model of Sterilizer: |
Problem Title:
DELTA 8 10 sterilizer Error message FAIL |
Problem Description:
DELTA 8 10 sterilizer Error message FAIL |
Reasons |
1. DELTA 8 10 sterilizer Error message FAIL The FAIL alarm is a response from serveral different sensors and aborts the cycle because a condition has been detected that will not allow the completion of a successful cycle. A) The air purging cycle (between 101C - 110C) takes more than 7 minutes B) The unit takes more than 45 minutes to reach sterilization parameters C) During the Sterilizatin portion of the cycle, the chamber temperature or pressure drops below minimum parameters for 3 minutes (accumulated time) D) The door circuit is interrupted after the Fill portion of a cycle is complete. |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |