Sterilizer Manufacturer:
Pelton & Crane
Model of Sterilizer:
Problem Title:
DELTA 8 10 sterilizer Error message FAIL - Unit fails between 110C and Sterilization Target
Problem Description:
DELTA 8 10 sterilizer Error message FAIL - Unit fails between 110C and Sterilization Target
Reasons
1. DELTA 8 10 sterilizer Error message FAIL - Unit fails between 110C and Sterilization Target

If unit takes too long time (more than 45 minutes) to get sterilization parameters then the sterilizer gets this kind of failure.
2. First of all, assume that the sterilizer can pass 90C - 110C air purging process successfully.

Then if the sterilizer can build pressure or not. If not, that means there is a leakage. Check H2O case for reference: how to find out leakage
3. If there is no leakage in the sterilizer, then we need to check its RTD steam sensor first, to make sure its reading is correct. To check this RTD sensor, just check its Ohm value matches real temperature or not. If not, replace it