DELTA 8 10 sterilizer does not dry properly
Problem Description:
DELTA 8 10 sterilizer does not dry properly
Reasons
1. DELTA 8 10 sterilizer does not dry properly

There are many reasons for this problem:

1) sterilizer is properly leveled or not?
2) how long the sterilizer was set for DRY
3) is the load properly packed?
4) is the air filter for the air pump dirty or clogged
5) is the air pump running during the DRY (to make sure the door is closed, if the door is open, the pump will be off)
6) test the air flow by removing the inlet of the filter
2. Sometime the checking valve on the rear of the sterilizer will leak, which will cause the filter and air pump out of work. For more details on air pump, check the following URL