Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: Tuttnauer 2540EHS |
Problem Title:
Tuttnauer Heat Up Problem |
Problem Description:
Low Heat Message is displayed |
Reasons |
1. Low Heat-- Refers to the temperature in the Chamber before sterilization has begun.This message is displayed if the programmed sterilization temperature is not reached within 50 minutes of the start of the instrument cycle or 80 minutes for the liquid cycle. |
Solutions |
1. Clogged Air Jet Very low voltage delaying heat up Bad Heating Elements or heater connection Safety Thermostat opening prematurely, turning off Heating Elements - only on units with control proms earlier than T93N3 Air Outlet Valve stuck close |
|
|||||