Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer Pressure Problem |
Problem Description:
Unit heats but pressure reading is inaccurate |
Solutions |
1. Check for a loose Pressure Transducer connection going into the back of the Ajunc board Check for a blockage in the copper tubing and silicone tubing going to the Pressure Transducer Turn off the power and disconnect and then reconnect all cables and connectors inside and outside the Electronic Box making sure that there are no bent or broken pins and all connections are solid. Recalibrate the Pressure Transducer Replace the Pressure Transducer and calibrate Replace the Ajunc Board and calibrate Follow the instructions regarding Control Problems |
|
|||||