Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: All of 'E' models |
Problem Title:
Tuttnauer 2340E model won |
Problem Description:
Low Pressure message is displayed |
Reasons |
1. Refers to pressure in the Chamber after the sterilization temperature has been reached. If the Chamber pressure drops 4psi below the pressure required for sterilization the cycle will abort. |
2. |
3. |
Solutions |
1. Insufficient water in the Chamber 2. Water leaking out of the Chamber 3. The sterilization phase of the cycle has been set for too long a period of time, allowing the Chamber water to boil away, and the chamber to run dry. 4. Heating Elements not cycling on and off properly. a. Problem with the Heat Solid State Relay (SSR) b. Problem with the control circuit 5. Bad Heating Elements -- not producing enough wattage 6. Safety Thermostat opening prematurely, turning off Heating elements, only on units with control proms earlier than T93N3 7. Bad Pressure Transducer |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |