Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: All of 'E' models |
Problem Title:
Tuttnauer 2340E models Pressure Problem |
Problem Description:
High Pressure message is displayed |
Reasons |
1. High Pressure-- Refers to Chamber pressure after sterilization temperature has been reached. If the Chamber pressure rises 10psi above the pressure required for sterilization the cycle will abort. |
Solutions |
1. 1. Sterilization temperature has been set above 273 2. Heating Elements are not centered or tightened properly. 3. Heating Elements remaining on instead of cycling on and off. 4. Pressure or Temperature calibration is incorrect Turn off the power and disconnect and then reconnect all cables and connectors inside and outside the Electronic Box making sure that there are no bent or broken pins and all connections are solid. Verify that the pressure and temperature reading are correct by using a mechanical pressure gauge and an independent thermometer. If the displayed pressure is not accurate then recalibrate the Pressure Sensor If the displayed temperature is not accurate then recalibrate the Temperature Sensor |
|
|||||