Tuttnauer
Model of Sterilizer:
Problem Title:
Tuttnauer E/EA/EK/EKA Overheating Problem
Problem Description:
Damage occurs inside the chamber during the sterilization cycle
Reasons
1. Either load unproperly or the unit is out of order

Solutions

1. Items are lying up against the Chamber wall. During sterilization the Chamber wall becomes very hot, much hotter than the steam temperature. A hot Chamber wall can damage anything that touches it.

The Chamber is not pitched properly

The unit is not filling with the correct amount of water

Clean the Water Sensor
Clean the Chamber Strainer on units manufactured after 2/2000
Clean the Pump Strainer on units manufactured after 2/2000
Reset the automatic water fill

The Water Sensor is being shorted by something metal (rack, trays or instruments), a damp cloth or bag, clear any debris away from the sensor

The amount of sterile time is set to high

The sterilization temperature is set to high

The items being sterilized are of an absorbent material and soaking up all the available steam

The unit has a steam or water leak

The Air Outlet Valve is stuck.

When the valve is stuck closed the Chamber may not fill with the proper amount of water.

When the valve is stuck open steam will escape to rapidly and the Chamber will run dry.

The Heating Elements are shorted allowing them to run continuously

The Heat SSR is defective allowing the Heating Elements to run continuously

The Safety Thermostat and / or Cut-Out Thermostat are not functioning properly. The probes maybe loose or miss positioned or the tubes maybe kinked.

The technical information is just to help sterilizer users as reference to do their trouble shooting.