Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer circuit breaker or fuse problem |
Problem Description:
Circuit breaker or fuse trips while unit is running |
Reasons |
1. If the main problem is Power Supply or Dry Pump fuse, please check them first |
Solutions |
1. Check that the voltage is in the correct range for that unit. The proper voltage range for 110 volt machines is between 110 and 25. The proper range for 220 volt machines is between 220 and 235 volts. Check the resistance of the Heating Elements, of particular interest would be a reading to ground. Check when the unit is hot but not plugged in Check for damaged or shorted wires throughout the unit |
|
|||||