Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer circuit breaker or fuse problem |
Problem Description:
Tuttnauer E/EA/EK/EKA circuit breaker or fuse trips as soon as unit is plugged in |
Reasons |
1. If the problem is with Power Supply fuse, please check the Power Supply fuse. |
2. If the problem is with Dry Pump fuse, please check with Dry Pump fuse issue |
Solutions |
1. Make sure this problem occurs when the units main switch is not turned on Unplug the unit Disconnect the wires coming out from the Circuit Breaker or Fuse Plug the unit back in If the problem persists then replace Fuse Holder or Circuit Breaker If the problem goes away then Check for a ground short at the Heating Elements, Check if the Thermostat capillary tubes are shorting the autoclave receptacle. Check for damaged or shorted wires throughout the unit |
|
|||||