Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer Dry Pump Problem |
Problem Description:
Dry pump blows fuse |
Reasons |
1. Pump is stuck or dry pump SSR out of work |
Solutions |
1. Check if the Dry Pump fan blade is stuck Check for a pinched wire or other ground short in the power leads going from the Dry Fuse to the Dry Pump. Turn off the power and disconnect the wires from terminals 1 & 2 on the Dry SSR. Turn the power back on , if the fuse blows again the problem is with the Dry SSR and it should be replaced. If the fuse does not blow then the problem is with the Dry Pump and it should be replaced. |
|
|||||