Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer E/EA/EK/EAK Door Problem |
Problem Description:
Door cover rubbing |
Reasons |
1. see solutions |
Solutions |
1. Determine the point at which the cover is rubbing or binding Determine which of the four mounting points is closest to the problem Once a mounting point has been isolated remove the screw from that point and insert from one to three flat washers between the metal tab of the door and the plastic mount of the door cover. Reinstall the screw. This procedure should push the cover out and eliminate the rubbing. If the problem is located an equal distance between two mounting points then both screws should be removed and washers inserted under both tabs. |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |