Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Air Outlet Valve Problem Tuttnauer E/EA/EK/EAK |
Problem Description:
Air Outlet blows off during sterile cycle |
Reasons |
1. some settings or readings are wrong |
Solutions |
1. Check that the temperature setting is not set any higher then 273F (134C) Check that the Heating Elements are turning on and off properly Check that the Heating Elements are installed properly |
2. Recalibrate the temperature sensor |
3. Recalibrate the temperature sensor (pressure transducer) |
|
|||||