Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Display Problem Tuttnauer E/EA/EK/EAK |
Problem Description:
Tuttnauer Display back light is very faint |
Reasons |
1. Either the power supply or the LED itself has problem |
Solutions |
1. Check the DC power between TP17 and TP1 to make sure a +5V there, otherwise, change the power supply board Use another LED to test. |
|
|||||