| Sterilizer Manufacturer: Tuttnauer |
| Model of Sterilizer: |
|
Problem Title:
Tuttnauer E/EA/EK/EAK Add Water Indicator Problem |
|
Problem Description:
Indicator is always on |
| Reasons |
| 1. see the solutions for detail |
|
Solutions |
| 1. Check that the Float Switch is free to move up and down Check the connector at JP2 on the Ajunc board for loose connection or a broken wire. Check between TP8 and TP1 for a DC voltage of 0 to 1 volt if the water level in the reservoir is acceptable and 3.5 to 5 volts if the level is too low. If the readings are correct but the indicator light remains on all the time then the problem is with the Digital Predg board or the Flat Cable between the Predg and Ajunc boards. Inspect the Flat Cable for loose connectors and bent or broken pins. Insure a good connection when reconnecting the Flat Cable If the problem is still not corrected then replace the Digital Predg board. If the readings are not correct then check continuity across the Float Switch. Cut the wires going to the Float Switch if necessary, connect an ohm meter and move the Float Switch up and down. When in the up position the meter should show continuity. When in the down position the meter should show an open circuit. If the continuity check fails then replace the Float Switch If the Float Switch checks out okay then there is a Control Problem |
|
|
|||||
|
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |