Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Cut-Out Theremostat Problems of Tuttnauer E/EA/EK/EAK |
Problem Description:
Cut-Out Thermostat does not turn off power to the Autoclave |
Reasons |
1. the good sensing probe should be installed in the upper channel of the rear most heating element, and tight it properly |
Solutions |
1. Check the sensor probe located properly and tightly |
2. Replace the cut off thermostat |
3. Check the heating element no a ground short bypass. |
|
|||||