Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Water Pump Problem of Tuttnauer E/EA/EK/EAK |
Problem Description:
Pump does not turn on |
Reasons |
1. It means now DC power supply to the water pump or the water pump is out of order |
Solutions |
1. Check the Water Pump fuse |
2. Check the connections between the water pump |
3. Check the Water Pump SSR (solid state relay) to make sure it works properly between the testing points TP13 and TP1, there is 10-12V DC for OFF and 0-1V DC for ON. If the Water Pump SSR does not work, then replace it, otherwise replace the Ajunc board |
4. Check the pump itself, it may be out of order |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |