Water Pump Problem of Tuttnauer E/EA/EK/EAK
Problem Description:
Pump does not turn off
Reasons
1. the DC power supply to the water pump keeps on

Solutions

1. Check the Water Supply SSR, to make sure 10-12V DC for ON, and 0-1V for OFF. If not replace the SSR
2. If the water pump SSR (solid state reply) works properly, the the problem is from the Ajunc board, which needs to be replaced.