Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Water Pump Problem of Tuttnauer E/EA/EK/EAK |
Problem Description:
Pump does not turn off |
Reasons |
1. the DC power supply to the water pump keeps on |
Solutions |
1. Check the Water Supply SSR, to make sure 10-12V DC for ON, and 0-1V for OFF. If not replace the SSR |
2. If the water pump SSR (solid state reply) works properly, the the problem is from the Ajunc board, which needs to be replaced. |
|
|||||