Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Water Pump Problem of Tuttnauer E/EA/EK/EAK |
Problem Description:
Water Pump blows fuse |
Reasons |
1. Some electric short related with the pump |
Solutions |
1. Check the pinched wire or ground short in the power leads going from the water pump fuse to the water pump Turn off the power and discoonect the wires from terminals 1 & 2 on the water pump SSR. Turn the pwoer back on, if the fuse blows again the problem is with the water pump or the pump capacitor and both should be replaced |
|
|||||