How to convert COX temperature reading between Fahrenheit and Celsius |
|
Problem Description:
The unit is reading 190 please tell how to get it back to Fahrenheit |
| Reasons |
| 1. This may need to be done usually after a power disturbance that affects the AC power to the sterilizer. |
|
Solutions |
| 1. To switch display between degrees F and degrees C: When the sterilizer is running: 1. Put it into shut-down-mode 2. While in shut-down-mode, press cycle 1 and 3 for a second at the same time. 3. Put sterilizer in start-up mode and display should now read the opposite of what it was, (centigrade or Farenheit). If it does not, unplug and plug sterilizer from AC. 4. If it still shows the wrong range, repeat steps 1-3 above. |
|
|
|||||
|