Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer 1730 EK Failure |
Problem Description:
Tuttnauer 1730 EK Failure |
Reasons |
1. Tuttnauer 1730 EK machine always failed in the Liquid Cycle. There is a 2 minutes dry time in the Liquid cycle settings. |
2. For some reason there is a 2 minutes Dry time for the liquid cycle. This will cause the machine always failed in the Liquid cycle. To fix this problem, just re-set the machine back to the factory default settings. |
3. I need the service manual of the sterilizer to tuttnauer, I know where I can find in the InterNet for download? |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |