Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer 2349 EKA no heat |
Problem Description:
Tuttnauer 2349 EKA no heat |
Reasons |
1. Tuttnauer 2349 EKA, after press the Start button, no heat up |
2. In most cases, if the display panel works well, but no heat up from the heater units, you need to check the SSR (Solid State Solenoid) ohm value, to make sure the SSR working well. However, before you check this SSR, which need you discounect power and remove the SSR (which is not easy to remove), it is better to check all the power wire connection first. We just met a case, the problem was from the loose connection of the power wires. Fixed by phone call with the cusotmer. |
|
|||||