Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: Statim 2000 |
Problem Title:
Statim 2000 Error 19 |
Problem Description:
Fault #19, the steam generator calibrationis invalid. A new calibration is required. How do I do this. |
Reasons |
1. A calibration needed |
Solutions |
1. 1) Turn off the main power 2) Remove the top cover 3) Short together the Pins W1 4) Turn on the power 5) LCD shows CALIBRATING Two digits on Upper Left, two digits on the Upper Right The Lower Left shows the temperature difference between boiler and chamber The Lower Right shows the chamber temperature 6) Insert cassette 7) Hold down the UNWRAPPED button and press START 8) an asterisk * will show after CALIBRATING 9) Takes about 5 minutes, a long beep will sound 10) Press STOP to vent 11) After LED shows PRESS STOP TO RESET, press STOP again 12) You get a new message SELECT A CYCLE 13) All set You may get a copy of technical manual for reference. |
|
|||||