Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
High Temp is displayed |
Problem Description:
High Temp is displayed |
Reasons |
1. "High Temp" is displayed High temperature and low pressure Steam trap is leaking: fix the leakage Needle valve is leaking: fix the leakage Temperature sensor PT100 is faulty: replace the sensor Pressure sensor is faulty: replace the sensor Temperature or pressure reading out of calibration: recalibrate Exhaust lines, chamber or jacket are leaking: fix the leakage Faulty steam trap: replace the trap needle valve open: close it completely then open half a turn |
|
|||||