Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: Statim 2000 |
Problem Title:
Scican Statim 2000 -error 16 |
Problem Description:
We get error 16 and the cycle will not go above 97 degrees. then it stops. I am in the middle of a clinic trying to do 35 surgeries a day. Please help. |
Reasons |
1. Either the water pump or pump filter out of work, or the boiler out of calibration |
Solutions |
1. Check the water pump and its filter first. If both of them are working well, then you need to recalibration the boiler. The problem is its temperature reading is wrong |
2. FYI visit Statim 2000 calibration |
|
|||||