Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: 2540 & 3870 EHS |
Problem Title:
Tuttnauer EHS "Low vacuum" is displaed |
Problem Description:
Low vacuum |
Reasons |
1. Tuttnauer EHS "Low vacuum" is displaed Chamber is leaking: fix the leakage Faulty vacuum pump: replace the pump Vacuum piping is leaking: fix the pipes Non-return (one way) valve is stuck: replace the valve Vacuum valve (52) does not work, maybe: faulty coil or plunger: replace the parts pump disconnected: check its wires faulty digital board DIG-T2: replace it faulty analogue board ANL-T2: replace it Air inlet valve (43) leaks or open Fast exhaust valve (73) leaks or open slow exhaust valve (74) leaks or open steam valve (93) leaks or open |
|
|||||