Sterilizer Manufacturer:
Tuttnauer
Model of Sterilizer:
Tuttnauer 1730 E/EK
Problem Title:
Tuttnauer - 1730 E - the chamber does not fill correctly only twelve seconds
Problem Description:
the chamber does not fill correctly we teach it but always fill twelve secons and that is not enough to run a cycle if you run it all the water goes to the water reservoir.

Thanks


Miguel Gomez
Reasons
1. The electrode may be dirty or defective.

Solutions

1. Please give us the serial no and year of the machine of this 1730E/EK(before 1993 or after). We want to know whether there is a water pump to fill the water or by gravity only.

You can check it by another way: Turn on the machine, there is a short moment to show the model of this machine. Check there is a WP at the end of the model or not. WP means Water Pump.

It seems this machine has 2 problems

Problem 1: No enough water filled inside the chamber
Problme 2: Some leaking, hard to say where. Maybe the solenoid valves, maybe door gasekt, maybe air jet by the safety valve, ... etc.
2. Try cleaning the electrode in the back/bottom of the chamber using scotch-brite or similiar.

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers