Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer EHS no response when the system is turned on |
Problem Description:
Tuttnauer EHS no response when the system is turned on |
Reasons |
1. Tuttnauer EHS no response when the system is turned on with AC but no DC check the connection for the DC supply check the Test Point on the ANL-T2 board to make sure both 5V and 12V DC supplies are OK Check a good grounding for both ANL-T2 and DIG-T2 boards: make sure the grounding screws are tighted properly |
|
|||||