Problem Title:
STATIM 5000 cycle fault
Problem Description:
I have been getting this message saying "cycle fault, not sterile 1". No error code is showing up. Not sure what to do. I have been having a hard time finding someone to service my machine. I'm in Ottawa, Ontario, Canada

Solutions

1. Normally it will not show "cycle fault, not sterile 1"
There were probably show "cycle fault #1". If this is the case, then mostly the problem is the heating element does not work. The machine got time out after a long time.
check the heating element ohm, it should be 11 ohma about.
check the fuse, make sure it works
check the power supply
make sure not too much load for the cycle.
You may need a technical manual to do further diagnosis

The technical information is just to help sterilizer users as reference to do their trouble shooting.