Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer EHS vacuum valve (52) always open |
Problem Description:
Tuttnauer EHS vacuum valve (52) always open |
Reasons |
1. Tuttnauer EHS vacuum valve (52) always open (3870EHS has 2 valves) Check the valve ground short connection at JP10-4 Check the DIG-T2 board TP8 voltage TP8<1V the valve will open Check the solenoid on the valve, if the voltage of coil is over 10V, then the ANL-T2 board is faulty |
|
|||||