Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: Statim 2000 |
Problem Title:
error code 26- not sterile |
Problem Description:
error code 26 keeps coming up and inst. are not getting sterilized machine is sitting level on counter top, and no kinks in the water hose. |
Reasons |
1. The chamber and steam generator (boiler) temperature were 2C degree apart fro at least 10 seconds during sterilization |
Solutions |
1. End user repair: check the exhaust seal screen, clean or replace as needed. |
2. Field Service Tech Repair: CAlbirate boilder. If the problem persists, replace and calibrate boilder |
|
|||||