Model of Sterilizer:
Statim 2000
Problem Title:
error code 26- not sterile
Problem Description:
error code 26 keeps coming up and inst. are not getting sterilized
machine is sitting level on counter top, and no kinks in the water hose.
Reasons
1. The chamber and steam generator (boiler) temperature were 2C degree apart fro at least 10 seconds during sterilization

Solutions

1. End user repair: check the exhaust seal screen, clean or replace as needed.
2. Field Service Tech Repair: CAlbirate boilder. If the problem persists, replace and calibrate boilder

The technical information is just to help sterilizer users as reference to do their trouble shooting.