Model of Sterilizer:
Problem Title:
Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up
Problem Description:
Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up
Reasons
1. Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up

1) check the temperature sensor connection to the JP2 on the ANL-T2 board
2) Check the PT100 sensor property: 32F should be 100 ohms, then increase 0.2 ohms for each degree F
3) use the a PT100 simulator to do test

The technical information is just to help sterilizer users as reference to do their trouble shooting.