Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up |
Problem Description:
Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up |
Reasons |
1. Tuttnauer EHS the reading temperature is higher than 284F or T1CUT showed up 1) check the temperature sensor connection to the JP2 on the ANL-T2 board 2) Check the PT100 sensor property: 32F should be 100 ohms, then increase 0.2 ohms for each degree F 3) use the a PT100 simulator to do test |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |