Sterilizer Manufacturer: Wayne Sterilizer |
Model of Sterilizer: Wayne Model 1000 |
Problem Title:
proper setting |
Problem Description:
What is the proper setting? i have the side dial on 8 and the timer to 70 mins. |
Reasons |
1. Call us for phoe help |
Solutions |
1. The side dial is used to control the temperature. Check the temperature reading on the sterilizer. For dry sterilizer, it is suggested to run at 375F. Depends on the load, for your timer settings. However 70 minutes is a little too long. We are hard to give you a suggestion on the minutes. What we can suggest you is to make sure run a dry indicator strip for each cycle. It will tell you the cycle is good or not. For more details, please visit the dry indicator strip site at following. |
|
|||||