Sterilizer Manufacturer: Midmark |
Model of Sterilizer: M11 |
Problem Title:
midmark model 9 intermittent error code 5 i have removed the bellows and it is clear any help would be appreciated |
Problem Description:
i keep receiving error code 5 intermittently and the bellows is clear |
Reasons |
1. The machine shuts down during sterilization cycle |
2. sterilizer overloaded |
3. bellows malfunctioning (closing early) |
4. sterilizer potentionmeter out of calibration and/or dirty temperature sensor |
5. temperature sensor malfunctioning |
Solutions |
1. Check that pressure vessel is not overloaded with heavy linen packs. instruct operator to reduce load size |
2. Bellows should close at approx 212F (100C). Replace bellows |
3. Clean temperature sensor, then check calibration of using pressure gauge test harness. Perform temperature/pressure potentionmeter adjustment. |
4. check calibration replace temperature sensor assembly and recalibrate. |
|
|||||