Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer EHS float switch in the reservoir is always ON or OFF |
Problem Description:
Tuttnauer EHS float switch in the reservoir is always ON or OFF |
Reasons |
1. Tuttnauer EHS float switch in the reservoir is always ON or OFF 1) make sure it can move free inside the reservoir 2) check all the wire connection to the float 3) check the ohm reading of the float in On/Off position to make sure getting different ohm reading 4) If the above is OK, then either ANL-T2 or DIG-T2 board is out of work |
|
|||||