Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: 2540 & 3870 EHS |
Problem Title:
Tuttnauer EHS water pump does not operate |
Problem Description:
Tuttnauer EHS water pump does not operate |
Reasons |
1. Tuttnauer EHS water pump does not operate 1) Mineral free water supply 2) Fuse 3) Check the TP17 on DIG-T2 board to make sure voltage is not <1V constantly, otherwise the DIG-T2 board is out of work 4) Check the pump power supply JP7 to make sure the JP7-2 on the AC-T1 baord is 120VAC otherwise the AC-T1 board is out of work 5) Check the pump mechanism or its piping system 6) the Lower float in the vacuum pump can stop the pump too |
|
|||||