Sterilizer Manufacturer: Pelton & Crane |
Model of Sterilizer: |
Problem Title:
TCR4C |
Problem Description:
Having trouble wiring the TCR4C for a OCR |
Reasons |
1. There is one more connector on the replacement TCR4C relay that needs a connection. |
Solutions |
1. Connect the two wires coming from the temp control on the front of the sterilizer to number 7 and 8 on the relay. You will need to cut the wire from the temp control to the timer...it will no longer connect to the timer. Connect number 4 and 5 to the thermistor and 123 connect the same as the old relay. |
|
|||||